FixedTEST Fri Dec 14 09:08:46 2018 Adrien MattaOTHERSocialN/ATest Post
MUGAST ELOG is now ONLINE
FixedTEST Wed Jan 16 16:33:28 2019 MarlèneGRITGeneral3-alphaStatus of detectors on 16/01/19
MG1 has some issue with the HT cable (for the BT bias). There is a shortcut only under vacuum. We checked the connector but we have to do vacuum again to check it.
Attachment 1: Detectors_status_160119.xlsx
Attachment 2: Capture_d’eÌcran_2019-01-16_aÌ€_17.05.57.png
Capture_d’eÌcran_2019-01-16_aÌ€_17.05.57.png
FixedTEST Wed Jan 16 17:36:30 2019 Valérian-MarlèneGRITGeneralN/ACalimero tests - How to configure Calimero in GECO
Tests of Calimero time calibration with MUVI1 channel 2 (MM2)

1- channel 3 of all mates selected (see screeshots)
2- channel 1 of all mates
3- channel 1 of all mates - with channels enabled !!!!!
4- channel 1 of all mates - with channels enabled !!!!!
5- all channels telescope MM2

Number of counts = total number of counts for one channel on all mates

Conclusion :
Options to be chosen
1- with channels enabled
2- 10000 for the statistics


Remarks
To run calimero on MUVI1, caliWatcher should be set to 10202 and CaliSlave to 10223
The opposite for MUVI2, caliWatcher 10223, CaliSlave 10202
Attachment 1: Calimero_MUVI1_MM2_voie3_allmates_bidim.png
Calimero_MUVI1_MM2_voie3_allmates_bidim.png
Attachment 2: Calimero_MUVI1_MM2_voie1_allmates_bidim.png
Calimero_MUVI1_MM2_voie1_allmates_bidim.png
Attachment 3: Calimero_MUVI1_MM3_voie1_allmates_bidim_withchannelsenabled.png
Calimero_MUVI1_MM3_voie1_allmates_bidim_withchannelsenabled.png
Attachment 4: Calimero_MUVI1_MM3_voie3_allmates_bidim_withchannelsenabled.png
Calimero_MUVI1_MM3_voie3_allmates_bidim_withchannelsenabled.png
Attachment 5: Calimero_MUVI1_MM2_allchannels_allmates_bidim_stat100000_withchannelsenabled.png
Calimero_MUVI1_MM2_allchannels_allmates_bidim_stat100000_withchannelsenabled.png
FixedTEST Fri Jan 18 15:32:50 2019 Adrien ValérianGRITGeneralN/ACalimero tests Muvi2 MM7 MG10C
After disabling nptool Multithreading, online is working fine again.

Concerning Calimero, it is currently working both on MUVI1 and MUVI2 for the X at least.

Joined here is the result of Calimero for MG10C on MUVI2 channel 3 (MM7).
Attachment 1: Calimero_MUVI2_MM7_allchannels_allmates__stat100000_withchannelsenabled.png
Calimero_MUVI2_MM7_allchannels_allmates__stat100000_withchannelsenabled.png
Attachment 2: Calimero_MUVI2_MM7_allchannels_allmates_withchannelsenabled.png
Calimero_MUVI2_MM7_allchannels_allmates_withchannelsenabled.png
Fixed  Tue Jan 22 11:15:15 2019 MarlèneGRITGeneralN/ASquare detectors
MG9C (thin square detector coupled with thick square detector unplugged) noisy
On X channels --> no noise problem ( about 100 ORD in the dark)
On Y channels --> MATE 10 & 11 noisy ! (see picture below with noisy Y piedestals)

With internal pulser (trigered by CK_TST) the counting rate is OK only if the thresholds are set at 1 MeV. If they are set at 500 keV, the counting rate is 10 times higher than the CK_TST

For MG10C, ORD is OK !
Attachment 1: MG9C_disabled_channels__no_crazy_ORD.png
MG9C_disabled_channels__no_crazy_ORD.png
Attachment 2: MG9C_disabled_channels__no_crazy_ORD_moredetailed.png
MG9C_disabled_channels__no_crazy_ORD_moredetailed.png
Attachment 3: MG9C_piedestal_of_disabled_channels_Y.png
MG9C_piedestal_of_disabled_channels_Y.png
OngoingTEST Tue Jan 22 18:52:23 2019 Valérian-MarlèneGRITGeneral3-alpha3 alpha source run on MG10C & MG3T
3 alpha source (not centered)
MG10C and MG3T ON
MG10C is connected to MG5T HT and BT and seen on MUVI1 channel2 -> online name=MM2
MG3T is connected to MG3T HT and BT and seen on MUVI1 channel3 -> online name=MM3

Attached : current and bias of the detectors
FixedTEST Wed Jan 23 09:11:26 2019 Valérian-MarlèneGRITGeneral3-alphaRUN9 : 3 alpha source run on MG10C & MG3T
RUN 9 (over night 22/01 ->23/01)
3 alpha source (not centered)
MG10C and MG3T ON
MG10C is connected to MG5T HT and BT and seen on MUVI1 channel2 -> online name=MM2
MG3T is connected to MG3T HT and BT and seen on MUVI1 channel3 -> online name=MM3

Attached : current and bias of the detectors (decreasing with time)
+ online spectra for 3 alpha source run

MG10C has 7 missing strips ! (see attached spectrum)
Attachment 1: 3alpha_sourceMG10C-MM2_MG3T-MM3.png
3alpha_sourceMG10C-MM2_MG3T-MM3.png
Attachment 2: HV_squares_plugged_on_MG4T-MG5T_andMG3T.png
HV_squares_plugged_on_MG4T-MG5T_andMG3T.png
Attachment 3: HV_squares_plugged_on_MG4T-MG5T_andMG3T_afternightrun.png
HV_squares_plugged_on_MG4T-MG5T_andMG3T_afternightrun.png
Attachment 4: MG10C_7missingstrips.png
MG10C_7missingstrips.png
Fixed  Wed Jan 23 12:24:10 2019 MarlèneGRITGeneral Change cable MG1T - Exchange of detectors named MG9C & MG10C
23/01/19

Change of the HT/BT cable of MG1T (the one that had a short cut before when the chamber was closed)

MG9C is now the square detector nb 4 (Micron number) and coupled with the thick detector --> this detector was working properly
MG10C is detector nb 3 (Micron number) --> this detector was noisy before
Fixed  Fri Jan 25 17:02:25 2019 MarlèneGRITGeneral Detectors numbering
Attachment 1: MUGAST_detectors.pdf
MUGAST_detectors.pdf
FixedTEST Thu Feb 7 09:52:08 2019 Freddy, MarlèneRUNGeneral3-alphaRUN10 : 3 alpha source on MG1-3-4-5-7
Test run with alpha source in front of trapezoid detectors.

First run with new LABELS updated from "MM" to "MG0" in GECO.

name : run_0010.dat.06-02-19_19h08m11s (1.4 GB)

date: 06/02/2019
start: 19:08
stop: ~22:00
Attachment 1: alpha_source_5trap_rawhistos_2.png
alpha_source_5trap_rawhistos_2.png
Attachment 2: alpha_source_5trap_rawhistos.png
alpha_source_5trap_rawhistos.png
Fixed  Thu Feb 14 08:50:58 2019 MarleneGRITGeneral3-alphaRUN 27
RUN 27 on 14th of feb.
3 alpha source (one in front of trap., one in front of square detectors)

MG10, MG4, MG7 and MG5 are OK.
MG1 shows alpha but at lower energy and with bad resolution (particularly on X). We suspect it is not well depleted. We tried to increase the bias up to 200V but did not see any change.
It is due to a non working SA/SL external cable that produces this error (we tested it with several detectors, we also tried to change the MUVI channel and it still does the same)

Conclusion :
We have tested all trapezoid detectors (MG1, MG3, MG4, MG5, MG7) the annular (MG11) and the thin square (MG10) they are all OK




Detectors (order in the GMT) :
MG10
MG1
MG4
-
MG7
MG5
-
-

attached : bias of detectors ( be careful : MG10 bias = MG11A)
spectra for all detectors
Attachment 1: HV_detectors_MG10_MG1_MG4_MG7_MG5.png
HV_detectors_MG10_MG1_MG4_MG7_MG5.png
Attachment 2: GMT8config_HV_detectors_MG10_MG1_MG4_MG7_MG5.png
GMT8config_HV_detectors_MG10_MG1_MG4_MG7_MG5.png
Attachment 3: 3alpha_sourceMG07X.png
3alpha_sourceMG07X.png
Attachment 4: 3alpha_sourceMG07Y.png
3alpha_sourceMG07Y.png
Attachment 5: 3alpha_sourceMG05XY.png
3alpha_sourceMG05XY.png
Attachment 6: 3alpha_sourceMG10XY.png
3alpha_sourceMG10XY.png
Attachment 7: 3alpha_sourceMG04XY.png
3alpha_sourceMG04XY.png
Attachment 8: 3alpha_sourceMG01XY.png
3alpha_sourceMG01XY.png
Fixed  Thu Feb 14 15:30:14 2019 Marlene Daniele JennyGRITGeneral3-alphaRUN 28
RUN 28 14th Feb
3 alpha source on MG11A, MG3T, MG1T

GMT channels
MG11
-
MG3
-
-
MG1
-
Attachment 1: 3alpha_sourceMG03XY._OKpng
Attachment 2: 3alpha_sourceMG11XY.png
3alpha_sourceMG11XY.png
Attachment 3: 3alpha_sourceMG01XY_OK.png
3alpha_sourceMG01XY_OK.png
Fixed  Fri Feb 15 15:31:28 2019 MarlèneGRITGeneralN/ADetectors bias
Fixedcom2019 Mon Feb 18 10:41:36 2019 MarleneGRITGeneralN/ABias & current of detectors
Attachment 1: bias.png
bias.png
FixedTEST Thu Feb 21 09:08:49 2019 FreddyRUNGeneral3-alphaRUN 30 - alpha on MG 1,3,5,7
Run 30

Alpha source run for trapezoid detectors during the night with pedestals aligned and zero suppression. Cooling at 15°C.
Thresholds at 500 keV.
Run size: 1.3 G

On MUVI1:

1--> MG5 (150V)
2--> --
3--> MG3 (160V)
4--> --

On MUVI2:
5--> MG7 (120V)
6--> MG1 (140V)
7--> --
8--> --

TRIGGERS are:
GMT ch1 --> OR MUVI1
GMT ch2 --> OR MUVI2
GMT ch3 --> OR MUVI3 (nothing on it)

Plots are now represented using the correspondance map which means that it correspond to the geometrical ordering on the detector. With this representation groups of strips missing or disfunctionning on a given region of the detector is more obvious.
Attachment 1: 3alpha_source_MG_1357_20Feb.png
3alpha_source_MG_1357_20Feb.png
Attachment 2: LowV_bis_run30.png
LowV_bis_run30.png
Attachment 3: LowV_run30.png
LowV_run30.png
Attachment 4: HV_run30.png
HV_run30.png
FixedTEST Thu Feb 21 14:37:26 2019 iulian, valerian &FreddyGRITHardware3-alphaResults Calibration 3-alpha RUN30 MG1, MG3,MG5 & MG7
Results of the calibration run 30
In order:
Telescope1 X,Y (MG5 X,Y)
Telescope4 X,Y (MG3 X,Y)
Telescope7 X,Y (MG7 X,Y)
Telescope9 X,Y (MG1 X,Y)
Attachment 1: Telescope1Summary.pdf
Telescope1Summary.pdf
Attachment 2: Telescope1Summary.pdf
Telescope1Summary.pdf
Attachment 3: Telescope4Summary.pdf
Telescope4Summary.pdf
Attachment 4: Telescope4Summary.pdf
Telescope4Summary.pdf
Attachment 5: Telescope7Summary.pdf
Telescope7Summary.pdf
Attachment 6: Telescope7Summary.pdf
Telescope7Summary.pdf
Attachment 7: Telescope9Summary.pdf
Telescope9Summary.pdf
Attachment 8: Telescope9Summary.pdf
Telescope9Summary.pdf
FixedTEST Thu Feb 21 16:43:45 2019 Valérian, Freddy, Iulian and MarlèneGRITHardwareN/ACurrent measurements
Joined are the the I(V).

Detectors seems to respond accordingly with the previous measurements from Marlène.

Annular detector curve is quite strange…

Joined files :

Depletion_Mugast_MG.pdf -> summary for trapezoidal + annular detectors
Depletion_Mugast_M2.pdf -> summary for Must2 detectors

Depletion_Mugast.xlsx -> Excel file
Attachment 1: Depletion_Mugast_MG.pdf
Depletion_Mugast_MG.pdf Depletion_Mugast_MG.pdf Depletion_Mugast_MG.pdf
Attachment 2: Depletion_Mugast_M2.pdf
Depletion_Mugast_M2.pdf Depletion_Mugast_M2.pdf
Attachment 3: Depletion_Mugast.xlsx
FixedTEST Thu Feb 21 17:32:42 2019 iulian, freddyGRITHardware3-alpharun 33 3 alfa with MG1,MG3,MG4,MG5,MG7,MG9 short run
First test of the repaired 2xcables. The repaired cables were placed between MUVI1_2 <=> MG4 & MUVI1_4 <=>MG11 (longer cable)
Correspondance for the rest is the same as for run 30. The mapping is the same for all detectors (including anulaire detector).

Alpha source run for trapezoid detectors + anulaire detector with pedestals aligned and zero suppression. Cooling at 15°C.
Thresholds at 500 keV.

On MUVI1:

1--> MG5 (150V)
2--> MG4 (95V)
3--> MG3 (160V)
4--> MG11 (60V)

On MUVI2:
5--> MG7 (120V)
6--> MG1 (140V)
7--> --
8--> --

TRIGGERS are:
GMT ch1 --> OR MUVI1
GMT ch2 --> OR MUVI2
GMT ch3 --> OR MUVI3 (nothing on it)
Attachment 1: 3alpha_sourceMG1MG4MG7MG9XY_run33.png
3alpha_sourceMG1MG4MG7MG9XY_run33.png
Attachment 2: 3alpha_sourceMG3MG5XY_run33.png
3alpha_sourceMG3MG5XY_run33.png
FixedTEST Thu Feb 21 18:14:48 2019 iulianGRITHardware3-alphaACQ problem run 34 3 alfa with MG1,MG3,MG4,MG5,MG7,MG9

Alpha source run for trapezoid detectors + anulaire detector with pedestals aligned (with HV Frown ) and zero suppression. Cooling at 15°C.
Thresholds at 500 keV.

On MUVI1:

1--> MG5 (150V)
2--> MG4 (95V)
3--> MG3 (160V)
4--> MG11 (60V)

On MUVI2:
5--> MG7 (120V)
6--> MG1 (140V)
7--> --
8--> --

TRIGGERS are:
GMT ch1 --> OR MUVI1
GMT ch2 --> OR MUVI2
GMT ch3 --> OR MUVI3 (nothing on it)

PROBLEM ACQ , MUVI2 has no validations (DECS). MUVI1 OK!. Looks like a problem on the catchslave actor of Narval as no events seems to pass trough

The problem was caused by the cable used to triger the GMTSlave from the FanIn FanOut.
Attachment 1: acq_problem_21and22fevrier.png
acq_problem_21and22fevrier.png
FixedTEST Fri Feb 22 11:56:53 2019 iulian GRITHardware3-alpharun 38 3-alpha with reordering of the MUGAST cables for the logic and analogic signals
Run 38

Alpha source run for trapezoid detectors during the night with pedestals aligned (with HV on) and zero suppression. Cooling at 15°C.
Thresholds at 500 keV.

Important:
The MUGAST signal cables (logic & analogic) have been reordered:


On MUVI1
ChannelSpectra LabelDetectorVoltage
1 MG1 MG1 140V
2 MG3 MG3 160V
3 MG4 MG4 95V
4 MG5 MG5 150V



On MUVI2:
ChannelSpectra LabelDetectorVoltage
1 MG7 MG7 120V
2 MG9 MG1160V
3 MG10 -- --
4 MG11 -- --


TRIGGERS are:
GMT ch1 --> OR MUVI1
GMT ch2 --> OR MUVI2
GMT ch3 --> OR MUVI3 (nothing on it)
Attachment 1: run38_3alpha_mg1345.png
run38_3alpha_mg1345.png
Attachment 2: run38_3alpha_mg711.png
run38_3alpha_mg711.png